erratic pH meter
Posted: 08 May 2010 09:57 AM   [ Ignore ]
Indispensable
Avatar
RankRankRankRankRank
Total Posts:  1450
Joined  2008-05-14

I was making a Colby this morning, and had calibrated my pH meter prior to use.  I did a 2 point calibration - to both 7.01 and 4.01.  During the cook, I was monitoring the pH which was steadily dropping.  It reached 6.15 and I was anticipating a 6.0 whereupon I would drain the whey.  However, the next check I did was at 6.26.  This seemed rather odd to me, so I recalibrated to find that the meter had shifted rather dramatically.  After recalibration, I checked the same sample and got a 5.85.  Now is this a somewhat normal phenomenon for a meter to go off so quickly; or does it appear that I may have done something else that affected it this way?  I think I caught it quickly enough that my cheese was not adversely affected; but only time will tell on that.

 Signature 

Rich

Profile
 
 
Posted: 09 May 2010 07:59 AM   [ Ignore ]   [ # 1 ]
Major Contributor
RankRankRank
Total Posts:  168
Joined  2009-02-02

I don’t think using a PH meter is as easy as we’re lead to believe. I too had problems yesterday. Making an Edam, start PH at 6.4 immediately after calibration, part way through it dropped to 6.3, then almost immediately went back to 6.5?
Until I have time to do more experiments I can’t accept these readings. I think my start PH was too low.

Profile
 
 
Posted: 09 May 2010 12:43 PM   [ Ignore ]   [ # 2 ]
Indispensable
Avatar
RankRankRankRankRank
Total Posts:  2415
Joined  2007-01-15

If the prob has not been pre soaked then it will drift. If the probe is older then a year, it is worn out even if its not used. trry the pre soaking (several hours) if that has not rectified it then contact dealer.

 Signature 

The Cheese Hole

Profile
 
 
Posted: 09 May 2010 07:08 PM   [ Ignore ]   [ # 3 ]
Indispensable
Avatar
RankRankRankRankRank
Total Posts:  1450
Joined  2008-05-14

I store the meter with the tip in water, so pre-soaking is not an issue.  Guess I’ll monitor it closely for the next couple makes and see what happens.  Its only 3 months old, so I’m fairly confident its not worn out.

 Signature 

Rich

Profile
 
 
Posted: 09 May 2010 09:01 PM   [ Ignore ]   [ # 4 ]
Major Contributor
RankRankRank
Total Posts:  168
Joined  2009-02-02
Green Cheese Maker - 10 May 2010 12:08 AM

I store the meter with the tip in water, so pre-soaking is not an issue.  Guess I’ll monitor it closely for the next couple makes and see what happens.  Its only 3 months old, so I’m fairly confident its not worn out.

Same here except this is only the second time I’ve used it. Mine is stored with ‘Storage solution’ in the cap, so mine hasn’t dried out either.
Is water a suitable storage medium?

Profile
 
 
Posted: 10 May 2010 08:35 AM   [ Ignore ]   [ # 5 ]
Indispensable
Avatar
RankRankRankRankRank
Total Posts:  2415
Joined  2007-01-15

Depends on the water, some tap water is full of cloreen and other corrosive substance. I use 7ph solution, distilled water would be better then tap water, or filtered water.

 Signature 

The Cheese Hole

Profile
 
 
Posted: 11 May 2010 07:54 AM   [ Ignore ]   [ # 6 ]
Active Contributor
RankRank
Total Posts:  30
Joined  2009-03-20

No, water is no good - in many cases it will damage the electrode. In order of chemically best use either KCl storage solution or pH 7 buffer solution.

Profile
 
 
Posted: 17 June 2010 07:01 AM   [ Ignore ]   [ # 7 ]
Major Contributor
Avatar
RankRankRank
Total Posts:  102
Joined  2008-03-04

I spent some time running lab analysis (PH and TA) at a local winery. I learned that your buffer solutions should be replaced about every 6 months along with the PH meter probe. I use a $50 handheld meter at home for wine and cheesemaking and am finding it inadequate. I just have a hard time trusting the readings even after calibration. As far as the buffer solutions go I would buy the small foil packets instead of the bottles so you can always have fresh solutions to calibrate with. I’ve come to the conclusion that if you really want to be sure of your PH readings you need a decent benchtop model starting at around $300.

Profile
 
 
Posted: 17 June 2010 09:17 AM   [ Ignore ]   [ # 8 ]
Indispensable
Avatar
RankRankRankRankRank
Total Posts:  2415
Joined  2007-01-15

Wow, not good news :(

 Signature 

The Cheese Hole

Profile
 
 
Posted: 14 July 2010 12:51 PM   [ Ignore ]   [ # 9 ]
Power User
Avatar
RankRankRankRank
Total Posts:  326
Joined  2008-03-21

Been away a while, but thinking of making another batch.  Looking at meters, and tried to find the phh7x, but seems to be discontinued.  Has anyone seen this meter:  HANNA HI 98103 Digital Water pH Meter Tester, Checker. Would this be a decent choice?  Now, when storing the probe, do you put it into a bottle of buffer solution or exactly what do you do for storage of the probe?  Thanks for the help.  Kim

 Signature 

Kim   cool smile

Profile
 
 
Posted: 14 July 2010 05:03 PM   [ Ignore ]   [ # 10 ]
Indispensable
Avatar
RankRankRankRankRank
Total Posts:  1450
Joined  2008-05-14

My Hanna Checker comes with a rubber probe cover that enables me to keep buffer solution around the probe tip during storage.

 Signature 

Rich

Profile
 
 
Posted: 14 July 2010 09:53 PM   [ Ignore ]   [ # 11 ]
Indispensable
Avatar
RankRankRankRankRank
Total Posts:  2415
Joined  2007-01-15

The rubber boots still dry out, make sure u wet it the day before.